Digital Inductance/Capacitance Meter June 2018 NAN uF Display, Calibration, LCD Issue

411 5
ADVERTISEMENT
  • NaN uF display despite frequency change

    #1 21680800
    Martin Walker
    Anonymous  
  • ADVERTISEMENT
  • I2C LCD address mismatch likely causing NAN display

    #2 21680801
    Alan Winstanley
    Anonymous  
  • ADVERTISEMENT
  • Calibration is 91.07 uH, not pH

    #3 21680802
    Martin Walker
    Anonymous  
  • ADVERTISEMENT
  • 370 kHz reading suggests 1 nF input capacitance

    #4 21680803
    Martin Walker
    Anonymous  
  • ADVERTISEMENT
  • Floating-point squaring overflows to nan

    #5 21680804
    Martin Walker
    Anonymous  
  • EEPROM CF initialization fixed NAN display issue

    #6 21680805
    Martin Walker
    Anonymous  