logo elektroda
logo elektroda
X
logo elektroda

Beckman Wavetek DM27XT Multimeter Reads OL on All V-Ohm/Com Functions After Arc

mholin 387 0
ADVERTISEMENT
  • #1 21188687
    mholin
    Level 5  
    Beckman Industrial/Wavetek Digital Multimeter DM27XT

    Was testing to see if a bug zapper grid voltage (=< 2800VDC) had discharged yet. Set meter to highest range (1000 VDC) and probed across the grid. Heard an audible arc so voltage must have discharged into something in the front end of the meter. Now all functions that use the V-Ohm/Com input just read "OL" (over range). Even the continuity test with the probes shorted reads OL with no beep. Opened it up and found input resistor R31 47 ohms, completely open. Replaced with new resistor but meter still reads "OL". Anybody have any insight into what other component to check?
  • ADVERTISEMENT
ADVERTISEMENT