GE QH1200T3/CL Halogen Bulbs Exploding in High Vacuum Sputter Deposition System

318 9
Best answers LABEL_AI_GENERATED

Why are GE QH1200T3/CL halogen heat lamps exploding and occasionally melting the lead wires in a 10^-6 torr sputter deposition chamber?

The most likely cause is overheating in this particular machine: the bulbs work in other machines, but here they are being run hotter and for longer, and in a vacuum they lose convection cooling so they depend mostly on radiant cooling [#21664446][#21664447][#21664445] Try lowering the lamp voltage and/or adding more lamps to achieve the same heat load, and see whether the failures stop [#21664445][#21664447] If possible, measure bulb temperature directly, because the chamber may make them run much hotter than the process-point temperatures you already monitor [#21664445][#21664447] Also check for fingerprints or other contamination on the quartz, since halogen bulbs can fail from oils, and inspect the lamp sockets and leads for loose connections or arcing, which can overheat and melt the wires [#21664442][#21664441]
AI summary based on the discussion. May contain errors.
ADVERTISEMENT
  • Vacuum conditions may be overheating the halogen bulbs

    #1 21664439
    Cody Tappan
    Anonymous  
  • ADVERTISEMENT
  • #2 21664440
    Sarah Harris
    Anonymous  
  • TVS diode test for intermittent supply faults

    #3 21664441
    Stephen Van Buskirk
    Anonymous  
  • ADVERTISEMENT
  • Loose connections and fingerprints can cause halogen tube failure

    #4 21664442
    Sarah Harris
    Anonymous  
  • ADVERTISEMENT
  • Heat sag may weaken quartz bulbs before implosion

    #5 21664443
    Cody Tappan
    Anonymous  
  • Current overshoot and finger contamination suspected

    #6 21664444
    Cody Tappan
    Anonymous  
  • ADVERTISEMENT
  • Lower lamp voltage and monitor bulb temperature

    #7 21664445
    Steve Lawson
    Anonymous  
  • Machine-specific cause likely from contamination or handling

    #8 21664446
    Stephen Van Buskirk
    Anonymous  
  • Bulb overheating likely from longer high-power operation

    #9 21664447
    Cody Tappan
    Anonymous  
  • Preheating halogen filaments may prevent bulb fracture

    #10 21664448
    Mark Harrington
    Anonymous  

Topic summary

LABEL_AI_GENERATED
A set of 16 GE QH1200T3/CL halogen bulbs rated at 144V, 1200W, and 350°C are used in parallel as a heater bank within a high vacuum sputter deposition system operating at approximately 10^-6 torr. The bulbs are powered by a Sorrenson 160/158 power supply in voltage control mode at about 130V and 128A total, delivering roughly 1040W per bulb. The bulbs have been repeatedly exploding during operation, with occasional melted wires observed despite using 12 AWG nichrome-coated copper multi-stranded wire with fiberglass insulation rated above 500°C. Potential causes discussed include exposure of standard halogen bulbs to high vacuum conditions, loss of convection cooling leading to reliance on radiant cooling, and thermal stress on quartz envelopes causing weakening and rupture. Fingerprints or contamination on bulbs during installation may cause premature failure due to localized overheating. Loose electrical connections can cause arcing, filament vibration, and excessive heating, potentially leading to wire melting. Voltage overshoot is minimal, but current overshoot up to 20A occurs. The problem is specific to one machine where bulbs run hotter and longer than in others, suggesting thermal management and temperature monitoring as next steps. Suggestions include lowering operating voltage to reduce bulb temperature, adding more lamps to maintain heat output, and implementing wireless temperature probes to compare bulb temperatures in vacuum versus atmosphere. Preheating filaments to avoid thermal shock and condensation-related failures was also mentioned as a preventive measure.