How can I use a 0-10V PID output to vary power to a 5V, 2A heating element?
Use a fixed 5V supply and let the 0-10V PID drive a logic-level N-channel MOSFET as the series pass element; the thread concludes this is a better fit than trying to find a special 0-10V-controlled AC-DC supply [#21664975][#21664976] A 2N7000 is not suitable, and you should pick a MOSFET that is fully enhanced around 4.5-10V gate drive, with enough current and thermal margin for the load [#21664979][#21664983][#21664985] For the MOSFET ratings, one reply suggests at least 1.5-2× the maximum drain-source voltage, so about 20V Vdss or more for a 5V circuit, and enough dissipation margin for at least 10-20W [#21664988] For low loss when on, aim for very low Rds(on); one example estimate was about 0.05 ohm to keep the drop near 0.1V at 2A [#21664988] The gate can be driven directly from the PID output with a common return, and no gate resistors are needed in that arrangement [#21664981][#21664983]
  • #1 21664967
    Ethan H
  • Clarifying voltage-controlled current source requirement
    #2 21664968
    Steve Lawson
  • #3 21664969
    Steve Lawson
  • Use a fixed 2 A supply with VCR control
    #4 21664970
    Ethan H
  • #5 21664971
    Ethan H
  • Op-amp circuit recommended over single transistor version
    #6 21664972
    Steve Lawson
  • #7 21664973
    Steve Lawson
  • 5V regulation needs about 5.7V input headroom
    #8 21664974
    Steve Lawson
  • Logic-level MOSFET replaces NPN for PID heater drive
    #9 21664975
    Earl Albin
  • #10 21664976
    Steve Lawson
  • #11 21664977
    Earl Albin
  • #12 21664978
    Ethan H
  • #13 21664979
    Earl Albin
  • #14 21664980
    Ethan H
  • 40V MOSFETs should work without gate resistors
    #15 21664981
    Earl Albin
  • #16 21664982
    Steve Lawson
  • #17 21664983
    Earl Albin
  • Darlington divider resistor values calculated for 10V input
    #18 21664984
    Steve Lawson
  • #19 21664985
    Steve Lawson
  • #20 21664986
    Ethan H
  • #21 21664987
    Ethan H
  • MOSFET voltage, current, and Rds sizing guidance
    #22 21664988
    Steve Lawson
  • #23 21664989
    Earl Albin
  • #24 21664990
    Steve Lawson
  • #25 21664991
    Steve Lawson
  • Logic MOSFETs offer more design options
    #26 21664992
    Earl Albin
  • #27 21664993
    Steve Lawson
  • #28 21664994
    Ethan H
The discussion centers on finding a linear, open-frame AC-DC power supply with a 0-10V control input to provide a variable current output at approximately 5V, intended for a heating element in biological research. The user aims to control the heating element power (max 10W, 5V at 2A) via a PID controller outputting a 0-10V linear signal. Challenges include achieving stable voltage/current regulation given the heating element's variable resistance and the need for linear control rather than PWM. Suggestions include using a fixed 5V, 2A power supply (e.g., HAA512AG) combined with a voltage-controlled resistor (VCR) or a linear control stage. The discussion favors replacing bipolar transistors with logic-level N-channel MOSFETs operating in the linear (ohmic) region, driven directly by the 0-10V PID output for efficient analog control. Key MOSFET parameters discussed are Vdss (recommended ≥20V), Id (≥2A), low Rds(on) to minimize voltage drop and power dissipation, and gate drive considerations (no gate resistor needed). IR (International Rectifier) MOSFETs are recommended, with references to IRF series devices optimized for 10V gate drive. Calculations for resistor values in bipolar transistor base drive circuits are provided, but MOSFET solutions are preferred for simplicity and efficiency. The user plans to test several MOSFETs to finalize the design.
TL;DR: Need a 0–10 V–controlled heater at 2 A, 5 V? Use a logic‑level MOSFET; “Rds needs to be low enough.” [Elektroda, Steve Lawson, post #21664988] Why it matters: This turns any 0–10 V PID output into simple, safe power control for 5 V/10 W heating loads.
What am I actually trying to build here?
A voltage‑controlled power stage for a 5 V/10 W heater using a 0–10 V PID output. The PID adjusts gate voltage, and the MOSFET throttles current through the heating pad. You are not regulating constant current; the PID corrects temperature as resistance changes with heat. [Elektroda, Earl Albin, post #21664975]
Why prefer a logic‑level MOSFET over a BJT pass transistor?
Logic‑level MOSFETs fully enhance at ~4.5 V, waste less power at low Rds(on), and need no base divider. That simplifies wiring and reduces drop. As one expert notes, use a logic MOSFET and drive the gate directly from the PID. [Elektroda, Earl Albin, post #21664975]
Which MOSFET specs matter most (Vdss, Id, Rds(on))?
Pick Vdss ≥20 V for a 5 V system. Ensure Id ≥2 A continuous. Aim for Rds(on) ≈50 mΩ so the drop is ~0.1 V at 2 A (≈0.2 W). Lower Rds(on) cuts heat and improves control margin. “Rds needs to be low enough.” [Elektroda, Steve Lawson, post #21664988]
How do I wire the basic linear MOSFET heater driver?
  1. Connect heater between +5 V supply and MOSFET drain.
  2. Tie MOSFET source to supply return.
  3. Drive MOSFET gate with the PID’s 0–10 V output (common grounds). No gate resistor is required in this linear use. [Elektroda, Earl Albin, post #21664975]
Do I need an op‑amp to translate 0–10 V?
No. With a logic‑level MOSFET, the PID’s 0–10 V can drive the gate directly, provided the grounds are common. This avoids Vce headroom issues that arise with BJT/op‑amp solutions on a 5 V rail. [Elektroda, Earl Albin, post #21664975]
Can I regulate exactly 5 V to the heater with a linear BJT stage?
Not cleanly from a 5 V supply. A BJT pass path needs headroom; expect about 5.7 V minimum to deliver a regulated 5 V load. That makes BJT approaches awkward at a strict 5 V rail. [Elektroda, Steve Lawson, post #21664974]
Why isn’t constant‑current control the goal for a heater?
Heater resistance changes with temperature, so constant current would mis‑deliver power. Let the PID regulate temperature by varying power. Use the MOSFET as a low‑loss throttle and let the closed loop correct for resistance shifts. [Elektroda, Earl Albin, post #21664975]
How big can inrush or transient current be as the heater warms?
Cold resistance can be lower, so current can briefly exceed 2 A. A worked example cites up to about 5 A during warm‑up, with modest MOSFET dissipation if Rds(on) is low. Design margin protects the device. [Elektroda, Steve Lawson, post #21664988]
What’s a VCR (voltage‑controlled resistor) and is it needed here?
A VCR is a component whose resistance changes with applied control voltage. It was considered conceptually to keep ~5 V across the load, but the thread’s consensus uses a logic‑level MOSFET instead. [Elektroda, Steve Lawson, post #21664969]
Is the 2N7000 suitable for this job?
No. The 2N7000 is a small‑signal MOSFET and not appropriate for a 5 V, 2 A heater stage. Choose a power MOSFET rated for several amps with low Rds(on). [Elektroda, Earl Albin, post #21664979]
How do I size power dissipation and heat sinking?
Estimate worst‑case linear loss. Example: target Rds(on) 50 mΩ gives ~0.1 V at 2 A, or 0.2 W steady. Allow more margin for warm‑up and control overhead. Selecting devices that can handle 10–20 W provides safe headroom. [Elektroda, Steve Lawson, post #21664988]
Would PWM be more efficient than linear control?
Yes. PWM reduces device dissipation and improves efficiency. The thread focuses on a simple linear stage, but acknowledges PWM as the efficient next step if redesigning the system. [Elektroda, Earl Albin, post #21664975]
How were the 420 Ω and 820 Ω resistors for the BJT divider calculated?
They set the base at ~6.4 V when the PID is 10 V, assuming a Darlington with ~1.4 V Vbe and Hfe ≈1000–2000. Divider current is chosen large versus base current to reduce loading error. The worked math yields ~470 Ω and 820 Ω. [Elektroda, Steve Lawson, post #21664984]
What gate‑drive range gives good control headroom?
Select a MOSFET that reaches the required current with about 8–9 V gate drive. This preserves dynamic range across the 0–10 V PID output for precise throttling. [Elektroda, Steve Lawson, post #21664985]
Does my specific 5 V/2 A open‑frame supply need tweaks?
Some 5 V supplies can be trimmed above 5 V, which helps BJT headroom, but verify current rating at the higher setting. For the MOSFET approach, standard 5 V is fine with a logic‑level device. [Elektroda, Steve Lawson, post #21664974]
Any example parts or sources to start testing?
International Rectifier logic‑level 40 V parts are suggested as drop‑in candidates. Order a few variants and test thermal behavior in your setup to confirm margin and stability. “Try the IR part and then explore.” [Elektroda, Earl Albin, post #21664981]