My copy arrived today. I chose ANENG AOS02 because it was the cheapest and for the price of the basic set of PLN 258.97 I got the PRO version with additional test leads (theoretically it was a more expensive option). The order arrived within 10 days. I have software version 1.03.51.
I checked the multimeter with a voltage standard and it is perfect, the error is +/- 1 of the least significant digit. Resistance measurement, also "on point" with my "standard" resistors. The multimeter`s measurement bandwidth is typical of 1 kHz (here the MT8206 with a bandwidth of up to 150 kHz simply "rocked"). The software version reminds you to use the correct measurement terminals each time you switch from voltage measurement to current measurement.
The oscilloscope remembers the settings, but the multimeter does not, and the operating mode (oscilloscope/multimeter) is not remembered either.
Indeed, waveforms above 2 MHz "float" a bit, but an oscilloscope up to 5 MHz is quite useful and you can easily recognize the shape of the waveform. 10 MHz with sampling 48 MSa/s is, of course, fiction. The interface is responsive and quite intuitive to use. The chart is refreshed very quickly, nothing "cuts off". Unfortunately, the sensitivity of the input track is poor - with the x10 probe, the minimum sensitivity is 200 mV/div. (old Velleman HPS40 had 50 mV/div), maximum only 100 V/div. There`s nothing crazy about the time base either - from 100 ns/div to 20 s/div (roll mode from 250 ms/div). In practice, this means that you can conveniently observe waveforms with a frequency of up to 2 MHz and an amplitude of at least 1 V. The trigger level is scaled proportionally to the sensitivity
and it cannot be set to 0 V (for the highest sensitivity it is 10 mV in 20 mV steps) . After performing the calibration procedure, it can be set to 0, but there is approximately -3% offset relative to the indicated trigger level, i.e. for a sensitivity of 1 V/div. (10V visible on the screen), when setting the trigger level to e.g. 0V, in reality it will be set to -300 mV.
Here you can see the shift of the trigger point (set to 0V):

By the way - after saving the screen, the waveform disappeared from the screen even though the oscilloscope remained in RUN mode. For it to appear, the acquisition had to be stopped and resumed.