Repair of Aneng AN8009 meter - erroneous ohmmeter readings

Olkus 4674 34

TL;DR

  • The Aneng AN8009 multimeter showed wildly wrong ohmmeter readings, misreading a 100 ohm resistor as tens of kiloohms and even displaying nonsense with open or shorted probes.
  • Cleaning the range switch and its PCB tracks fixed the fault; oily grease had leaked from the switch onto the board, and the contacts were washed as well.
  • The meter’s original 100uF supply electrolytic was upgraded to 1000uF, one of the recommended AN8009 modifications.
  • After reassembly, the ohmmeter readings returned to correct values and the repair was completed in about half an hour at a cost of 0zł.
  • The rotary switch is a common failure point in many meters because it also handles power switching, which accelerates wear.
Generated by the language model.
ADVERTISEMENT
Treść została przetłumaczona polish » english Zobacz oryginalną wersję tematu
📢 Listen (AI):
  • #31 21222833
    398216 Usunięty
    Level 43  
    Posts: 34016
    Help: 3914
    Rate: 9190
    keseszel wrote:
    But I remember the Isostats siphoned off.
    .
    And what do the SILVER contacts of the Isostats have to do with the gold plated contacts of the range switch? Silver reacts very readily with sulphur (from the air and there in turn from exhaust fumes etc.) and is therefore coated with silver sulphide - a hard coating that is perfectly (in comparison) insulating . Gold is a metal that does not react with any chemicals in the air. It is therefore not exposed to any resulting problems.
  • ADVERTISEMENT
  • #32 21222893
    jajacek44
    Level 26  
    Posts: 714
    Help: 76
    Rate: 169
    @398216 Usunięty just that gold is soft and without lubrication it migrates causing short circuits .
    Unless it is "hard gold" .
  • ADVERTISEMENT
  • #33 21223096
    398216 Usunięty
    Level 43  
    Posts: 34016
    Help: 3914
    Rate: 9190
    jajacek44 wrote:
    gold is soft and without lubrication migrates causing short circuits
    .
    And silver does not.
    Both theses are worth the same amount. Especially the statement about migration. And for gilding, the so-called "technical gold" alloy based on gold is used, which is much more mechanically resistant.
    If we are talking about short circuits, it may be more about silver and more specifically about the problem of the occurrence of spontaneous micro-shorts in connections soldered with "silver tin", i.e. a tin-silver solder alloy.
  • ADVERTISEMENT
  • #34 21223207
    jajacek44
    Level 26  
    Posts: 714
    Help: 76
    Rate: 169
    Migration, i.e. the transfer through the gold contacts to the laminate .
    I don't think my colleague has had much contact with gold-plated connectors or switches.
  • #35 21224030
    CMS
    VIP Meritorious for electroda.pl
    Posts: 8440
    Help: 256
    Rate: 2581
    I confirm the occurrence of this phenomenon.
📢 Listen (AI):

Topic summary

✨ The discussion revolves around the repair of the Aneng AN8009 multimeter, specifically addressing erroneous ohmmeter readings. The user initially suspected battery issues but found them to be in good condition. Subsequent troubleshooting pointed to potential problems with grease on the contacts, which may have caused erratic readings. Various participants shared insights on the use of silicone grease versus petroleum jelly, with some suggesting that grease can introduce resistance and lead to inaccurate measurements. Others noted that dirt and corrosion could exacerbate contact issues. A participant resolved their problem by replacing the banana sockets with higher quality components, resulting in improved resistance readings. The conversation concluded with considerations on whether to lubricate contacts, particularly gold-plated ones, and the implications of using different types of lubricants.
Generated by the language model.

FAQ

TL;DR: If your Aneng AN8009 shows 100 Ω as tens of kΩ, clean the rotary switch PCB first. One user fixed it in ~30 minutes for 0 zł after saying the ohmmeter was "going crazy." This FAQ helps AN8009 owners restore correct resistance readings and check whether the real fault is the switch, sockets, or leads. [#21210334]

Why it matters: Low-ohm faults in the range switch can corrupt basic continuity and resistance checks, making an otherwise usable multimeter unreliable.

Option What users observed Cost / effort Best use
Clean switch tracks with IPA Restored correct ohmmeter readings after grease and dirt removal 0 zł, ~30 min First repair step
Re-grease after cleaning Some users report 1 year of stable readings with fresh grease Extra materials, repeat teardown If the original grease was contaminated
Leave switch dry Chosen by the repair author after cleaning 0 zł When gold-plated tracks already work correctly
Replace banana sockets / improve leads Reduced shorted-probe reading from 10–20 Ω to 0.3–0.4 Ω in one case ~25 PLN + 1 hour If switch cleaning does not fix low-ohm error

Key insight: The thread points to contaminated switch grease and dirty range-switch tracks as the main cause of wildly wrong ohmmeter readings. If cleaning does not restore near-zero readings with shorted probes, check the input sockets and test leads next.

Quick Facts

  • The original repair fixed the meter after cleaning the rotary switch PCB, took about half an hour, and cost 0 zł. [#21210334]
  • The failure symptom was severe: a 100 Ω resistor could read as several tens of kΩ, and shorted probes could show several hundred ohms. [#21210334]
  • The author had used the meter since the first half of 2021 on the same batteries, but battery replacement did not solve the problem. [#21210334]
  • One user improved low-resistance performance by replacing poor banana sockets, cutting shorted-probe readings from 10–20 Ω to 0.3–0.4 Ω; the mod cost about 25 PLN and 1 hour. [#21211496]
  • A common optional AN8009 modification mentioned in the thread is increasing the power-supply capacitor from 100 µF to 1000 µF while the meter is already open. [#21210334]

How do I fix an Aneng AN8009 that shows wildly wrong resistance values on the ohmmeter range?

Start by cleaning the rotary range switch PCB and its contact area. In the reported repair, batteries were ruled out first, then the meter was opened, oily dirt was removed from the switch tracks, and correct readings returned immediately. The whole job took about 30 minutes and cost 0 zł. If the meter still shows high resistance with shorted probes after cleaning, check the banana sockets and test leads next. [#21210334]

What symptoms point to a dirty or failing rotary range switch in a digital multimeter like the Aneng AN8009?

The clearest symptom is unstable resistance readings that change with switch movement. In the thread, a 100 Ω resistor read as several tens of kΩ, open probes showed nonsense values, and shorted probes could show several hundred ohms instead of near zero. Another strong clue is that switching ranges several times can briefly restore normal operation. [#21210334]

Why can switching ranges several times temporarily restore normal ohmmeter readings in an Aneng AN8009?

Repeatedly turning the selector can momentarily improve contact between the wiper and dirty PCB tracks. The author observed that normal readings sometimes returned after moving through several ranges, which strongly implicates the rotary switch rather than the batteries. That temporary recovery usually means contamination or wear is present, not that the fault is gone. [#21210334]

What is the proper step-by-step way to disassemble, clean, and reassemble the AN8009 range switch PCB?

Use a simple 3-step process. 1. Remove the case, take out the PCB, and expose the rotary switch tracks. 2. Clean the visible grease and dirt from the PCB tracks and the underside contact part; the author used IPA. 3. Reassemble the meter and verify with a known resistor and shorted probes. In the thread, this restored correct readings without replacing parts. [#21212730]

Which cleaner works best for removing leaked grease and dirt from multimeter switch tracks: IPA, Contact 61, or potentiometer spray?

IPA is the only cleaner in the thread that clearly solved the fault. The author answered directly that the switch was cleaned with IPA, and after that the AN8009 returned to correct ohmmeter readings. Contact 61 and potentiometer spray were only proposed later as possible options, not confirmed fixes in this repair. [#21212730]

Silicone grease vs non-acid technical petroleum jelly: which is better for rotary switch contacts in a multimeter?

The thread does not show a clear winner, but it leans toward silicone grease over petroleum jelly for relubrication. Several users discussed silicone grease as the intended type, and one user reported a year of stable operation after cleaning and applying fresh SGB grease. Petroleum jelly was raised as an option, but no successful multimeter result was documented for it in the thread. [#21212353]

How much does the contact material matter when choosing a lubricant for multimeter switch contacts?

It matters a lot, because users in the thread treated gold-plated and silver contacts as different cases. Gold-plated rotary tracks were debated because grease can either protect a thin surface or raise contact resistance in low-current measurement circuits. By contrast, silver Isostat contacts were mentioned as a separate case with different aging behavior. [#21222833]

What is SGB grease, and why do some users recommend it for rotary switch contacts after cleaning?

"SGB grease" is a contact lubricant that users apply to rotary switch interfaces after cleaning, with the key characteristic that it is used in a very thin replacement layer after old grease and dirt are removed. One user recommended it because an AN8009 with similar symptoms worked correctly for a full year after cleaning and regreasing with SGB. [#21212353]

What is an Isostat switch, and how is it different from the gold-plated rotary range switch used in meters like the AN8009?

"Isostat" is a pushbutton switch family that routes signals through silver contacts, a key characteristic that differs from the gold-plated circular PCB tracks used in a multimeter’s rotary range selector. In the thread, users stressed that old Isostat experience does not transfer directly, because silver and gold contacts age and react differently. [#21222833]

Why do some users say grease protects gold-plated switch tracks, while others say it increases contact resistance and ruins low-ohm readings?

Both views came from different failure modes seen in practice. Some users said grease reduces friction and wear on thin gold plating, especially in selectors that also act as power switches. Others reported that contaminated or misplaced grease raises contact resistance, and one commenter argued this is especially harmful in modern meters that measure resistance with very low current. [#21214752]

When should I leave a multimeter range switch dry instead of relubricating it after cleaning?

Leave it dry when cleaning alone restores stable readings and the gold-plated tracks look healthy. That is exactly what the repair author decided after the discussion, choosing to run the AN8009 without new grease and monitor long-term wear. This is the safest choice when the original grease appears dirty, migrated, or electrically troublesome. [#21215616]

How can dirty grease, metal filings, or corrosion on switch tracks create leakage paths and cause false resistance measurements?

Dirty grease can spread across adjacent switch fields and stop the contact from making a clean, low-resistance connection. One user specifically pointed to grease contaminated with micro filings or corrosion, saying it can begin to conduct to some extent and bridge contact areas. In an ohmmeter circuit, that leakage distorts low-value readings and can make shorted probes read far above zero. [#21210879]

What other faults besides the range switch can make an Aneng meter read 10–20 ohms with the probes shorted, such as bad banana sockets or poor test leads?

Bad banana sockets and poor leads can cause the same symptom. One user reported an Aneng that showed 10–20 Ω with the probes shorted, then traced the fault to cheap internal sockets made from tinned sheet material. The thread also warns that some cheap test leads can add several ohms on their own, so verify leads and jacks before blaming the ADC or firmware. [#21211496]

How do upgraded banana sockets and better test leads improve low-resistance readings in Aneng multimeters?

They reduce series resistance between the meter input and the probe tip. In the thread, replacing the original sockets with gold-plated bronze types changed the shorted-probe reading from 10–20 Ω down to about 0.3–0.4 Ω, and the modified meter also passed a 10 A load test. Better leads help for the same reason: less contact loss and less conductor resistance. [#21211496]

Why do users replace the Aneng AN8009 power-supply capacitor from 100 uF to 1000 uF, and what effect does that modification have?

Users do it as a recommended AN8009 modification while the meter is already open. The repair author replaced the original 100 µF electrolytic with a 1000 µF part during the same teardown, citing an existing modification thread. This change was not presented as the cause of the ohmmeter fix, but as a separate preventive or performance-oriented upgrade. [#21210334]
Generated by the language model.